In 2021 we tested an Alarm-Assisted Natural Ventilation system in a school located in Northern Italy.
The system was based on real time CO2 measurements and optimized UI with enhanced acoustics timely sending windows opening requests to students and teachers (a procedure also referred as Signalled Manual Airing). Our system had tunable CO2 multi-thresholds and displayed specific instructions to be followed for each different CO2 threshold level.
For instance:
1. IF CO2 > th1 = 700 ppm, students were asked to open one window for 10 min (triggering a low ACH)
2. when CO2 was still > th1 students were asked to open BOTH windows for 20 min (activating partial cross-ventilation flow from outside one window into the other one --> medium ACH)
3 when CO2 was >> th2 = 1500 ppm, students were asked to open BOTH windows + DOOR until CO2 was < th1 (activating cross-ventilation flow from both windows toward the open door to rapidly decrease the CO2 concentration ---> high ACH).
[ACH = air changes per hour]
In WINTER, after a learning period, we repeatedly achieved impressive results from students self-controlling the CO2 indoor levels: the 6h-averaged CO2 concentration was close to 1000 ppm (in a V = 135 m3 and N = 20 students+1 teacher) which correspond to an avg ACH between 5.5 and 6 h-1.
The attached graph shows this comparison: in green the experimentally measured CO2 concentration curve from assisted NV vs in red the theoretical concentration simulating MV with steady ACH = 5.5 h-1 (this curve is easily obtained solving the CO2 mass balance equation for the same contextual classroom data)
Issues to be discussed:
1) can alarm-assisted-NV achieve comparable MV performances (under specific circonstances (like a school located in a suff. "windy" region) ?
2) can hybrid systems (assisted NV combined with smaller and more cost-effective MV units) be an option to improve ventilation in schools ?
3) can alarm-assisted-NV due to his 5-8 times lower total costs be an option to improve ventilation in schools on a LARGE scale (millions of school buildings worldwide suffer of poor ventilation condition and no-budget to afford MV/HVAC systems)
PS: all data are taken from our recent Energy & Building 2024 publication
"Benefits and thermal limits of CO2-driven signaled windows opening in schools: an in-depth data-driven analysis"
Thank you in advance for your time!
I hope a constructive discussion can follow.
Alessandro
Mechanical ventilation plays a crucial role in the management of respiratory failure in the intensive care unit (ICU). Respiratory failure occurs when the respiratory system fails to maintain adequate gas exchange, leading to hypoxemia (low oxygen levels) and/or hypercapnia (high carbon dioxide levels) in the blood. Mechanical ventilation provides artificial support to the respiratory system, assisting or replacing spontaneous breathing to ensure adequate oxygenation and ventilation.
Mechanical ventilation primarily supports respiratory function, while an IABP provides temporary circulatory support in cardiac failure. ECMO, on the other hand, offers comprehensive cardiopulmonary support and can be tailored to provide respiratory, cardiac, or combined support in critically ill patients with severe respiratory or cardiac failure. Each modality has specific indications, limitations, and risks, and the choice of therapy depends on the underlying pathology, patient characteristics, and available resources.
I am conducting an exergoeconomic analysis of HVAC system energy consumption in a classroom, considering three parametric cases: air conditioning only, air conditioning with mechanical ventilation, and air conditioning with a membrane heat exchanger. I am struggling to determine the specific processes I should follow and employ for my research. Should I perform the analysis using software or manual calculations? This study is an extension of a previous study conducted by my senior, and he has provided me with the electricity consumption data for the three cases. Could you please provide some guidance?
This is an urgent request triggered by an actual situation. After an incidental pneumonectomy, a woman now 72 years old, suffering from longstanding COPD combined with a restrictive lung disease du to secondary deformations of the thoracic cage, remains fully conscious and mentally undisturbed while intubated and under mechanical ventilation. She is neither prepared to nor intends to die. Apparently, her respiratory center is highly dysfunctional but does not appear to be completely knocked out.
Would it be possible.
a) to stimulate the respiratory center by narrowly balancing the hypoxia and hypercapnia tolerance with special ventilator settings?
b) to directly stimulate the respiratory center?
According to the concept of one-size- doesn't fit all in anesthesia and particularly in mechanical ventilation, should we think again about the fixed upper normal level of driving pressure especially in morbidly obese patients and pediatrics ?
In these papers "Karla Lynch" is listed. The real name is "Kevin Lynch".
Accuracy in the measurement of endogenous nitric oxide in the mechanically ventilated patient.
September 2000
Anesthesiology 93(Supplement):A55
DOI:
10.1097/00000542-200009001-00551
Robert E. Black
H.A. Tillmann Hein
Michael A E Ramsay
Karla Lynch
Paradoxical Role of Inhaled Nitric Oxide in Advanced Liver Disease
Article
October 1999
Proceedings (Baylor University. Medical Center)
Robert I. Simpson
Michael A E Ramsay
Karla Lynch
[...]
In situation of anesthesia or need of mechanical ventilation in icu due to respiratory failure..
Ventilation during adult cardiopulmonary resuscitation (CPR) is poorly understood. Therefore, guideline recommendations are limited. The use of waveform capnography is in part recommended to monitor frequency. Other ventilation measurements such as tidal volume or inspiratory pressure are not regularly obtained, especially when a bag-valve system is used. The use of new monitoring devices can improve guideline adherence and could lead to better understanding of ventilation during CPR. Different EMS systems have varying levels of training, equipment and resources during CPR of out-of-hospital cardiac arrest (OHCA) patients. To better understand the current state of ventilation monitoring during OHCA CPR researcher/practitioner feedback and international perspectives on this question are needed and very much appreciated.
The study of ventilation during adult cardiac arrest remains challenging due to the unexpected nature of sudden cardiac arrest and the limited resources/personnel on site. This is especially true for interventions that influence outcomes when applied early in the cardiac arrest phase. Therefore, animal models (i.e. pigs, dogs), manikins, human cadavers and computer models have been used to study intra-arrest ventilation. Also, some data has been made available from registries and clinical studies in humans.
While the possible answers to my question heavily depend on the respective research question, personal perspectives on the well known experimental models, as well as lesser known models for this niche of cardiac arrest research, would be very much appreciated.
Please note, that I do not to intend to discuss airway management during cardiac arrest. Although, I'm aware that both intra-arrest ventilation and airway management are closely connected.
I am currently writing my dissertation on the use of non-invasive ventilation to deliver nitric oxide in neonates and I was wondering:
- What are people‘s experiences of using non-invasive iNO with CPAP, Nasal cannula, oxygen hood etc?
- Which gestational have you primarily used it with?
- What were the indications/ underlying pathologies?
- Have you found this has reduced the need for mechanical ventilation or ECMO?
- Have you needed to deliver higher doses to achieve the same effect seen on mechanical ventilation?
- Which countries have you seen this being practiced?
Any other insights or information would be greatly appreciated.
The preliminary report from the randomized RECOVERY clinical trial (NEJM JW Infect Dis Sep 2020 and N Engl J Med 2020 Jul 17; demonstrated that 10 days of dexamethasone resulted in a mortality benefit in hospitalized COVID-19 patients, especially those on mechanical ventilation. Investigators at a referral center in Brazil have now performed a double-blind, randomized, placebo-controlled clinical trial evaluating the efficacy of a 5-day course of methylprednisolone (MP) at reducing the mortality of patients hospitalized with COVID-19.
Of 416 patients randomized, 393 (mean age, 55 years) completed follow-up: 194 in the MP arm and 199 in the placebo arm. No patient received remdesivir, anti-IL-6, or anti-IL-1 agents. The most common comorbidities were diabetes, hypertension, and alcohol use disorder. One third of patients were mechanically ventilated. Mortality at day 28 was 37.1% in the MP group and 38.2% in the placebo group. No between-group differences were apparent in mortality at 7 days or 14 days, viral clearance in the upper airways, or need for mechanical ventilation at 7 days. In a subgroup analysis, day-28 mortality was significantly lower with MP versus placebo among patients older than 60 years of age (46.6% vs. 61.9%).
What is your thought on this ongoing debate? should we use it or not?
Hi fellow researchers and colleagues
I'm starting a new online peer review journal specific for the topic of mechanical ventilation. It will be free for authors and readers. Hoping for first issue in September 2020
Any suggestions how to promote it worldwide or feedback ?
Thank you
Ehab Daoud
Immune responses to infections with by a corona virus vary widely and are appear to be related to the development of most severe complication, acute respiratory distress syndrome. Since survival of patients respondingto the virus in this way depends on respirators support, mechanical ventilation and extracorporeal oxygenation, therapeutic methods which demand highly specialized medical and nursing staff, human resources which become scarce in an epidemic or pandemic. Since vaccination are not available in newly emerging corona virus epidemics it would be interesting to know if and which targeted pharmacological modulation of immune response early in the course of an infection could help to reduce the need for intensive care and/or improve the outcome of respiratory support.
As a consequence of the Noise Abatement Act, they were obligatory for many noise-loaded dwellings. Many types were developed.
Later, recently in fact, energy saving measures changed this: mechanical ventilation including heat exchangers and silencers made "suskasten" unnecessary.
Covid-19 cases, who are developing severe ARDS / respiratory failure are requiring mechanical ventilation, and the number of such cases is increasing during this pandemic. However, even in the countries having advanced and robust health care facilities are failing to provide an adequate number of ICU beds and Ventilators for such increasing numbers of patients. Healthcare providers of many areas are now using one ventilator for providing mechanical ventilation to multiple patients simultaneously. As most of the ventilators can not provide differential ventilation (even those which can, are designed for two lungs at most 2 patients), is it safe to do so? As the compliance of the lungs of the different patients will be different, their PS, Vt and MV required will be different, how it can be feasible? Isn't there a chance of cross-infection / super-infection? Please opine and guide..
The entire world is now affected and the resources are getting exhausted. Even the best healthcare systems are failing to tackle the explosion of COVID-19. Judicious use of resources is therefore very essential. Many of the COVID-19 patients will be frail, terminally ill, etc. in whom intubation/mechanical ventilation or resuscitation may be futile. So, is it high time to adopt DNR/DNI for such patients? If yes, for which patients? Should the administrations/governments/ethical aspects be less rigid so that such rules can be adopted? What is the local rule/policy in your place?
Anyone interested in exploring novel ways of monitoring the appropriate position of endotracheal tube (ETT) in mechanically ventilated ICU patients?
Traditionally this is done with repeated Chest X-rays. What about researching new available technologies, like ultrasound or wireless methods, without the potential hazards of repeated radiation?
This is the situation. A classification model with 5 groups with some kind of order has recently been published in adult patients under mechanical ventilation (http://bit.ly/33E3T2V) and I want to test if this classification also suits for pediatric patients in terms of similarity of proportions. I wonder if Chi2 (in a 2 x 5 table) vs. Cochran-Armitage test would be the best way to test this situation (I think the latter, because of the ordered classification) and also which would be the minimum sample size for each test. I found here (http://bit.ly/2OWB7Gz) that sample size in this scenario gets smaller when number of groups increases, and that with a five-group classification a n=400 would satisfy the requirements for a power near 80%.
Occasionally we used to have transcutaneous CO2 monitors for selected ICU patients, especially those who are ventilated with HFOV.
However, with its poor correlation with the blood gas analysis and the complications of the electrode site burns, I think it is out of fashion?
Is anyone using similar products with better clinical experience?
I don'twant to use CFD method.
Waiting to read this article after reading the excellent article 'Ventilating the newborn and child.
I am interested in any model of lung physiology that can help with studying modes of ventilation in newborns.
Please take part to this international survey (3 minutes!). We want to know how you manage anticoagulation, including antithrombin supplementation, during veno-venous ECMO. Your contribution will be acknowledged in case of publication (this is why the survey is not anonymous).
what is the best management in Acute exacerbation of COPD which is resistant to medical management :
A- CPAP ( Non invasive positve pressure ventilation )
B- Intubation and mechanical Ventilation .
We know the values of NOx, CO and soot production for vehicles based on the PIARC reports (for fluent traffic, as well as for traffic jams), the traffic flow from a detailed traffic forecast analysis and the age/fuel type of the cars as brackets from national statistics. This allows us to estimate mean/peak emissions within the tunnel with reasonable accuracy. We know what is the airflow velocity / ventilation capacity for all of the traffic scenarios as well. Now, we have to propose a filtration solution, for which we first need to know, how can we translate our emissions into PM2.5 / PM10.
A very easy assumption for our further modelling, would be that the most of the particulates generated in the tunnel are formed from the soot. I'm not sure if it is possible for larger particulates to form from NOx / O3, as the tunnel is constantly ventilated (either through traffic, or by mechanical ventilation in a traffic stop). I would also assume that mass of soot >>> mass of particulates emitted by breaks, tyres etc., which means we could ignore the latter (especially for traffic stop scenario).
Please share your experience in this (particulate) matter.
I am simulating a mechanical ventilation of a building floor space. Requesting help on how to define the Boundary Conditions (In general, I came across materials using P-Q curve of fans as inlet definition for electronic cooling applications. Is it similar for Jet fans/Propeller fans ? How should I do it inside FLUENT ? Thanks for your help in advance!
PS: Is there any sample *wbpz files that any one can share so that I can take a look the Named Selection definition ? This would be of great help!
Is there a positive effect on parallel canullamanegement and the weaning of ventilation? Or is there to less evidence of effectiveness?
Hello all,
In case of compressible fluids, where temperature can change due to changes in pressure, is there some threshold (or minimum) change in pressure required after which one can observe change in temperature. ( one may consider evaluating the phenomena on time scale of micro seconds)
Similarly, is the pressure change required to cause change in temperature in compression same as expansion i.e will the same change in pressure (increase or decrease) cause change in temperature or is it that for change in temperature in expansion, a little higher (or lower) change in pressure may be required as compared to compression. ( here also considering the processes being evaluated at micro second time scale)
Thanks
In my view the cause of lung damage during modern mechanical ventilation is due to the positive pressure. Positive pressure ventilation causes atelectasis, and subsequently all the other problems if mechanical ventilation is maintained for a long time.
Negative pressure (expanding the thorax and getting air to flow in by under-pressure e.g. by a quirass sytem, or natural breathing) takes away the atelectatis very rapidly. That is why in every manual of the Anesthesiogists it says: directly after surgery, when the patient is awake again, ask him/her to take e few deep sighs.
My question: is it possible by your thechnique to prove the development of atelectasis by positive pressure and removal of the atelectasis by negative pressure vetilation?
This is very important, because patients with severe lung problems should not be ventilated by positive pressure systems. The life of a category of patients will be saved by negative pressure ventilation, because of the above reasons. So, please show the world what is going on, for most of the ventilators still believe the "law" (spoken out as a proposition around 1900) that there can be no difference between negative and positive pressure; they claim that everything is determined only by the pressure difference. They claim this is just physics. But this is only true in a static situation. During dynamic pressure variation this is absolutely not true.
Jan van Egmond.
In respiratory research a breathing apparatus consisting of mouthpiece, filter, Pneumotachometer, and non-rebreathing valves plus some connectors are usually used. Although a non-rebreathing valve is used to reduce dead space, each of these devices has its own dead space. Though small, adding together they build a relatively large dead space sometimes. What is the max acceptable dead space in a breathing apparatus, for a study including healthy adults?
I’m working on project about the development of electronic communication device to be used with invasive mechanically ventilated patients who are not able to speak verbally, so beside I need to measure its impacts on patients outcomes , I am looking to measure the nurses experiences or satisfaction while using this new device.
How many times a week and for how many months should patients with Chronic LBP be trained in a hydrotherapy setup so as to improve core muscles, back extensor and gluteal muscle strength and endurance of core muscles and back extensors? Please provide any supporting articles as well if available.
Can the use of different types of mechanical ventilatory modes cause peripheral vasoconstriction?
During the cares of a patient on BIPAP ASB it was noted that the patients fingers were both cold on both hands, and mottled on the left hand. After changing the ventilatory mode to CPAP ASB, both hands and fingers warmed up, and return of normal colour appeared.
But on returning the patient back to BIPAP ASB, after an hour the patients fingers became cold again!
Has anyone got any specific reasoning as to why, if any, the reason for the peripheral vasoconstriction due to different modes of mechanical ventilation, be it increased Peak Pressures etc?
Many Intensivists straight away intubate and mechanically Ventilate the patients. In my personal experience > 100 instances majority of the patients in Myasthenic crisis recover completely with Inj. Neostigmin Slow IV bolus upto 1 to 1.5 mg over 5 minutes period without intubation. Soon they needed IM Neostigmin with oral Pyridostigmin. Some of these patients also received IV Ig. Rarely 2 - 3 % crisis patients needed intubation and Mechanical Ventilation.
We do intraoperative contrast enhanced ultrasounds of the liver with Sonovue. When reaching the liver there seem to be too many bubbles destroyed.
- We use the same technique/device on prostates (transrectal, in awake non ventilated patients in the outpatient clinic). This works fine.
- In the operating room we do them during liver resections before the resection phase. We pre-oxygenate patients and pause the mechanical ventilation for a second. We give a fast bolus of 2.4ml and a slow one (5-10 sec) in which we do a flash 3 times.
Any ideas or experience with the intraoperative use?
As lung protective strategy you prefer volume control or pressure control ventilator mode?!
During controlled mechanical ventilation as compared to spontaneous breathing, less gradient of pressure in the system exists helping with repetitively opening and closing of the alveoli. When lungs are ventilated e.g. using the positive pressure mode, by application of positive end expiratory pressure (PEEP) it prevents alveolar over distension during cycles by avoiding its repetitive opening and closing. The systemic venous blood return still depends on a pressure gradient between the extrathoracic veins and the RA (the right pressure gradient) to create adequate RV preload, but with lesser amount....(can this be quantified, and how?). AM I COMPLETELY WRONG by saying that: during controlled mechanical ventilation the inspiration does not significantly increase this gradient to the level as observed in case of spontaneous breathing to accelerate venous return while enhancing the preload?
Do you have any publications/books where I can find the information about duration of a single respiratory rehabilitation intervention on a mechanically ventilated patient in Intensive Care Unit?
Cardiff university adult nursing student
interested if there is any welsh studies or policies regarding this topic.
In need of direction of the best way to direct my literature review.
Patient has major stroke on the left middle cerebral artery, right side no sensation except the foot, with atrial fibrillation and polymyositis. Would the intercostal muscles be strong enough to allow breathing ? Patient is 79 year old female.
The guidelines recommend change each 72-96hs in the overall patient, and every 48 hours in patients with COPD, but the practice is very different between different hospitals
PTP is usually measured to determine the inspiratory effort of the patients with spontaneous breathing. The static recoil of the chest wall has to be added in the calculation. My question is how to get the static recoil of the chest wall? Or please provide a reference.
Whereas in the ninety’s blind nasotracheal intubation (NTI) was the gold standard for medical ICU’s patients (more than 90%; Vassal et al, Intensive Care Med 1993) and the surgical ICU’s patients suspected or requiring mechanical ventilation more than 48 H (Aebert et al Intensive care Med 1988), after the implementation of rapid sequence induction (RSI), NTI’s use became confidential (less than 1% in a recent survey). Therefore, NTI is no longer taught in the ICU’s, whereas it may be necessary in some particular cases (inability to open mouth, to move the neck…) and reduces at least the risk of unplanned extubation. Moreover RSI is not so safe and easy according to the recent meta-analysis of Hubble et al (Prehosp Emerg Care 2010).
So should we save nasotracheal intubation?
I am currently interviewing SCI patients for my research study and some report the lack of mouthcare in critical care as being particularly upsetting. This should be provided routinely as part of a daily care bundle, however these patients have had to ask for it - despite being non-verbal at that time due tracheostomy/vent.
I'd like to find out about practices elsewhere.
Is there a correlation between the number of days premature infant required mechanical ventilation increase chance of requiring or needing bronchodilator therapy?
Hi to all,
I construct a plant including a way for passing of air. I can measure pressure at the start and end of tube and flow rate and maybe temperature. I add an extra artificial orifice as resistance Intentionally. How can i calculate airway resistance of this combination with these measures ?
As you know, the relation between Pressure, flow and airway resistance are obtained as below : R = (P_end - P_start) / (Flow_av) where Flow_av=(Flow_end + Flow_start)/2.
1- I have done numerous experiments with different flow rate. In every flow rate, resistance is different but i expect the airway resistance (based on electrical analogous) is constant ! isn't it ?!
2- Also i have done numerous experiments with constant flow rate and different artificial orifice. When i add a orifice with R=5 cmH2O/lit/s, the airway resistance (obtained from above eq.) was 6 cmH2O/lit/s. Then I add another orifice with R=5 cmH2O/lit/s and repeated the experiment again, but this time the value of airway resistance was 7 cmH2O/lit/s. What is wrong ?!!
Performing tracer decay testing in non-well mixed spaces.
Why not use a MULTIVIB mattress for transferring sound stimuli to these patients?
It can transfer music as well at VAT stimuli, and will aid the process of ventilation significantly.
Olav Skille
Need responses for an article to be published in a respiratory care magazine.
Do you use heat moisture exchangers or active humidifiers? Do you know if there is a guideline or a suggested material and values for optimal results?
In acute myocarditis with severe LV dysfunction, worsening pulmonary edema leading to progressive hypoxemia requires mechanical ventilation. Higher Peep can help in improving oxygenation which in turn is beneficial for failing myocardium. On the other hand, it will decrease cardiac output in already compromised status. What should be the aim in such patients and is there any role of permissive hypocapnea with higher ventilation rate to maintain alkalotic environment?
Hi all, I am investigating on different ways to increase the efficiency of Iranian wind towers. Right now, I am doing a research on contemporary buildings having passive wind tower(or Cool tower) like Animal Campus Dog Adoption Park project design by tsk (http://www.aiatopten.org/node/154). I wonder if you know any other project like this.
Looking forward to hear from you.
Thanks in advance.
Are there any specific number of these procedures you need to perform under supervison to achieve competency? How many you need to do to maintain this competency?
For Mechanical ventilation in laparoscopic in Trendelemburg positioning.
Dual modes of ventilation seem to be more physiological.
In a patient under sedation and paralytic agent on mechanical ventilation, is there any difference between pressure control mode and SIMV PS mode? What would be the preferred mode?
Can anyone tell the negative airway pressure in the trachea during inhalation, during calm respiration but also in a forced deep inhalation, model human 60 - 80 kg?
In a standard shunt model, blood oxygen content is decreased by venous admixture. Saturated capillary blood mixes with blood from the shunt. Because of the nature of the blood dissociation curve, higher FiO2 and, hence, higher alveolar partial O2 pressure will increase saturation only slightly above a FiO2 of about 30%. Hence, it will only increase SaO2 slightly if the shunt fraction is constant.
So why does increasing FiO2 above 30% in a real patient still increase SaO2 efficiently? I thought it could be based on increased diffusion, but did not find any literature about it.
There are studies on higher mortality in burn patients with inhalation injury. Cancio 2004 found out, that inhalation injuries were linked to higher volumina of fluids applied, only when artificially ventilated. Could it be possible that the higher mortality of inhalation injuries at least partially is linked to the fact of hyperbaric ventilation, increasing the level of natriuretic peptide and causing capillary leakage in a higher amount than in non-ventilated, followed by higher needs of fluid?
I have heard of people using APRV, which seems a little odd in the absence of inspiratory efforts. My impression is that there is not much research on this topic. Any thoughts?
Positive end-expiratory pressure(PEEP) is usually applied during mechanical ventilation to improve lung compliance and oxygenation, and then elevate oxygenation index in patients with respiratory failure, meanwhile, PEEP could increase intrathoracic pressure and influence the accurate measurement of hemodynamic parameters. But current studies to explore the effect of PEEP on hemodynamics were mostly performed in patients with acute respiratory distress syndrome or with acute lung injury,and the results drawn from these studies were influenced by multiple factors. Our study is to determine the impact of PEEP on such hemodynamic parameters as central venous pressure(CVP), mean arterial pressure (MAP) and heart rate in patients with central respiratory failure in neurological ICU, aiming to supply some quantitative guide in the acurate evaluation of some hemodynamic parameter levels measured during PEEP application in mechanically ventilated patients.
Were you surprised by the two studies published in the NEJM, Oscar and Oscillate? Is the benefit of HFOV lost in the era of low volume, pressure limited ventilation for ARDS? Will you continue using HFOV for rescue therapy or jump to extracorporeal support instead? Why do you think Oscillate had a higher mortality in the HFOV group?
The number of patients under mechanical ventilation increases, but we have limited health professionals in our ICU so, it would be better to have the automated MV, to protect the patients.