Hello! I‘m a student and the topic of my research is developing a method for quantitive analysis of glucosamine and chondroitin (in tablets after dissolution) via HPLC-DAD.
We‘ve been using tosyl chloride for derivatisation and NaOH for hydrolysis of chondroitine. After analyzing glucosamine and chondroitin (after hydrolysis and derivatisation) separately and together, we got just one peak. It seems that those are derivatized glucosamine and galactosamine together.
We tried using gradient mode, changing pH and mobile phases, but it didn’t make things better at all. I also tried ion-paired additives (octanesulfonate), but I was told that it can destroy the column so I stopped. Is it even possible to separate these two compounds with a C18 column?
I would be very grateful for your advice!
Thank you in advance!
Relevant answer
Answer
I am not sure that RP chromatography is the way to go. HILIC might be something to try - or a mixed mode WAX/RP or WAX/HILIC. Imho, though, I would suggest HPAE-PAD - like Ismail already did. To boost this statement, please find a paper published by my colleagues linked below.
You find all the experimental data, the does and don'ts, described in great detail.
All the very best for your research
Detlef
  • asked a question related to Glucosamine
Question
1 answer
I am trying to separate glucosamine and N-acetyl glucosamine from a filter-sterilized biological sample containing salt, amino acids, and other impurities. I used a standard solution containing an equimolar (2.5mM) solution of glucosamine+ N-acetyl glucosamine in milli-Q water. However, I am only getting one single peak of N-acetyl glucosamine (retention time 2.5 minutes). Can anyone help me with the separation process? I tried the following conditions:
Mobile phase A: 10mM sodium acetate+10 mM sodium octanesulfonate in water at pH 5.1
Mobile phase B: MeOH
Injection volume: 5 µl
Flow rate: 0.7ml/min
Column temperature: ambient
Column: Nucleodur C18 Gravity, 250x4.6 mm, 3 µm
Relevant answer
Answer
You can't. Carbohydrates are simply too hydrophilic and don't interact with C18 columns. Either use another column which is more suitable for carbohydrates or look for a derivatization method that yields more hydrophobic analytes.
  • asked a question related to Glucosamine
Question
2 answers
I need how to made standard curve for   N acetyl D glucosamine in A582 you published it in paper "Chitinolytic assay of indigenous Trichoderma isolates collected from different geographical locations of Chhattisgarh in Central India"2012 
i need a method to prepare the dilution of NAGA with  1% DNS "3,5  dinitrosalycilic acid " 
how to prepare the NAGA ?? in this Method
Relevant answer
Answer
"0.1% (w/v) N-Acetyl-D-Glucosamine Standard Solution (NAG) (Prepare 10 ml in deionized water using N-Acetyl-D-Glucosamine"
  • asked a question related to Glucosamine
Question
3 answers
Hi,
I need help understanding qPCR analysis, i have calculated the mean CT, ∆Ct, ∆∆Ct and 2^-∆∆Ct is this called the livak method? Where can i see where if there is downregulated or upregulated?
I have also dona ANOVA multiple comparisons (tukeys) test on SPSS and i have a control and two treatment groups, where the significance is <.001 what does that mean? please help!! I need some guidance.
Relevant answer
Answer
You can call the method deltadeltaCT.
Once you calculated your fold change values for each sample, you can see which samples are higher and which samples are lower than 1, your reference sample. To know if the differences are significant, the first generic test is to calculate standard error based on your biological replicates and see if there is or not overlap. Other tests may be suitable depending on the setup and actual values, as indicated already.
  • asked a question related to Glucosamine
Question
2 answers
Gram-negative bacterial pathogens interact with the host cells by using type III secretion systems (T3SS) to inject virulence proteins into cells. NleB effectors are glycosyltransferases that modify host protein substrates with N-acetyl glucosamine (GlcNAc) on arginine residues. This post-translational modification disrupts the normal functioning of host immune response proteins. T3SS effectors are thought to be inactive within the bacterium and fold into their active conformations after they are injected, due to the activity of chaperones . El Qaidi, et al (2020) reported that the bacterial glutathione synthetase (GshB) is glycosylated by NleB resulted in enhanced GshB activity, leading to an increase in glutathione production, and promoted bacterial survival in oxidative stress conditions. Is it possible that true role of T3SS effectors in bacteria was overlooked?
El Qaidi, S., Scott, N.E., Hays, M.P. et al. An intra-bacterial activity for a T3SS effector. Sci Rep 10, 1073 (2020). https://doi.org/10.1038/s41598-020-58062-y
(4) (PDF) An intra-bacterial activity for a T3SS effector (researchgate.net)
Relevant answer
Answer
There are many examples of proteins and pathways that serve multiple roles. So it is certainly possible that the T3SS has multiple roles which include both stress response and virulence and these activities may be regulated differently as well. There is no real answer to which is the "true" role.
  • asked a question related to Glucosamine
Question
6 answers
Dear all,
I'm currently interested in the hexosamine-pathway and especially in the enzyme Glutamine fructose-6-phosphate amidotransferase (GFAT).
The end product of the enzymatic reaction of GFAT is D-glucosamine 6-phosphate. It is known that glucosamine can enter the cells when added to cell culture via glucose transporters, but there is no data in the literature whether D-glucosamine 6-phosphate can be transported by these transporters as well. Does somebody know whether D-glucosamine 6-phosphate can enter the cell through glucose-transporters or maybe via alternative ways? And my second question is, whether D-glucosamine 6-phosphate is stable in cell culture conditions at all?
Kind regards,
Robert
Relevant answer
Answer
Leonid Volkov and Takayuki Kitagawa here is a recent and well illustrated review on G6P exchange and G6Pase in endoplasmic reticulum
  • asked a question related to Glucosamine
Question
4 answers
Hi,
I have been trying on a USP method regarding Glucosamine Sulfate Potassium Chloride. The assay is rather simple. It requires the use of amino-phase column to be used in reversed phase mode (which some may call HILIC?). The column i am using is Phenomenex Luna NH2 column (150 x 4.6mm, 5micron). The mobile phase used is premixed 75% acetonitrile: 25% phosphate buffer.
After about 20 injections, i discovered that the retention time shifts forward (earlier elution). Initially, it was around 10min (as indicated in the monograph), later shifted to about 5min and stabilised. This is on of the issues. However, a second issue later arises. After about 140 injections, the theoretical plate numbers fell steadily from 2400 to 1200. Then, i tried washing the column as instructed by the manufacturer with EtOH, followed by Hexane for 20 column volume. Unfortunately, the efficiency fell sharply to 700 (i might have killed the column?).
Here are some of the things that i have been doing:
  1. I always filter and degas my mobile phase.
  2. The column is stored in 70% ACN after each run.
  3. There is a washing procedure at the end of each run with 5:95 AcN;water, followed by 100% THF then 95:5 AcN:water.
  4. We always use HPLC grade solvent and MilliQ water.
  5. The buffer was 20mM potassium phosphate dibasic (pH 7.5).
  6. We always condition the column for at least 100min before run.
  7. The pressure remain stable at 55bar from the beginning.
  8. I did flush the HPLC system with 50% IPA to eliminate the system contamination issue.
There are actually a few questions on my mind right now:
  1. I have read some discussion about amino phase column being used as reversed phase. Some of them mentioned that this kind of usage is prone to have column bleeding/deterioration problem. Can anyone share his/her experience about how true is this?
  2. Apparently the column is in trouble and losing efficiency. Can anyone share some knowledge on ways to save/salvage the column?
  3. There is peak broadening issue which i think is the main contributor to the loss of efficiency. In my case, i can only think of column bleeding/loss of bonded phase. Can it be some other causes? As i have not changed any of the HPLC parameters (flow rate, inj vol etc)
  4. After mixing of the pH7.5 buffer and acetonitrile, the pH rose to 8.8. Is this significant to the column lifetime? Phenomenex said the column can withstand up to pH11.
Your help and sharing is much appreciated! Please let me know if i have left out any info to allow you to better understand my situation.
Thanks!!
Relevant answer
Answer
(1) Replace the HPLC column with a new column.
(2) Incorporate flushing with IPA mixtures on your Amino column . IPA is very effective in washing the water and other materials off the surface of the support that THF and ACN do not. *Developing a proper column method with any column (esp an AMINO) insures the longest column lifetime, best %RSD and accurate results. Failure to wash and re-equilibrate a column correctly after each use is the most common reason for Rt and peak shape changes over time.
(3) Amino column are very fragile and many not last long.
  • asked a question related to Glucosamine
Question
6 answers
Please suggest me a standardized method to prepare DNS reagent for the determination of reducing sugar by using N acetyl D glucosamine.
Thank you
Relevant answer
Answer
Dear Renuka,
Preparation steps of 3,5-dinitrosalicylic acid solution 100(ml):
1. Dissolve 30(g) of Potassium sodium tartrate tetrahydrate in 20(mL) distilled water. Add the salt to water gradually.
2. Prepare 2(N) sodium hydroxide solution 20 (mL). Dissolve smoothly sodium hydroxide in water using a glass rod and stirring the mixture of granules in water.
3. Dissolve 1(g) of 3,5-dinitrosalicylic acid in 50 (mL) of distilled water while the solution is mixed by magnetic stirrer with hot plate at 90-95˚C.
4. Add gradually the solution of potassium sodium tartrate tetrahydrate (prepared solution in Step 1) to 3,5-dinitrosalicylic acid solution (prepared solution the Step 3) while the solution is mixed by magnetic stirrer with hot plate at 90-95˚C.
5. Add slowly 2(N) sodium hydroxide solution (the solution prepared in step 2) to the solution prepared in step 4 while the solution is mixed by magnetic stirrer with hot plate at 90-95˚C.
6. After the components are completely dissolved, filter the final solution by filter paper.
7. Transfer the solution in dark glass bottles and storage at ambient temperature.
good luck!
  • asked a question related to Glucosamine
Question
4 answers
Hi everyone. I am currently working on HPLC assay of glucosamine, in accordance to USP method. Below are some of the specs of the method:
  1. Phenomenex Luna L8 column, 4.6x150mm, 5microns
  2. Temperature: 35 degreeC
  3. Flow rate: 1.5mL/min
  4. UV detector wavelength: 195nm
  5. Injection volume: 10microlitre
  6. Mobile Phase: premixed (buffer:AcN=25:75)
There are several problems I am facing currently, such as the retention times are not at around 10min (as stated in USP). Also, I am having a problem of delaying of retention time after each injection. For example, a sequence of 6 injections that I have ran gave such retention times, from injection 1 to 6: 5.265, 5.279, 5.294, 5.303, 5.313, 5.322.
The equilibration of the column took about 2hour before the initiation of sequence.
Did anyone encounter such situation before? Your advice is much appreciated.
Relevant answer
Answer
Femida Patel Hi Femida. Thank you for your suggestion. I will try it out and see if it improves.
Udaya Jayasundara hi Udaya. Yes i prepared the mobile phase fresh prior to every run. In fact I also did try with one-day old mobile phase and it didn't affect much. The conditioning you mean is the equilibration before experiment starts? Your advice is much appreciated.
  • asked a question related to Glucosamine
Question
5 answers
I have been trying to synthesise it using Glucosamine HCl and Succinic Anhydride along with triethylamine. But I am unable to get a good yield of the product. Kindly provide valuable suggestions
Relevant answer
Answer
pl suggest solvent used in reaction .it should be a protic otherwise succinic anhydride will converted to succinic acid
  • asked a question related to Glucosamine
Question
3 answers
Osteoarthritis is characterised as the deterioration of the articular cartilage which then causes bone-to-bone contact. Glucosamine is a primary constituent of cartilage proteoglycans, which is why there are claims that glucosamine could possibly be an anti-arthritic. Despite this, there has been no current and impressive evidence supporting this claim. Has there been any current advances on this claim and if so, what are they?
Relevant answer
Answer
Many of the products used for arthritis like glucosamine etc are more of neutraceuticals than valid pharmaceutical agents. In early arthritis many may improve symptoms but we'll established cartilage degeneration is not repaired by any of drugs at present.
  • asked a question related to Glucosamine
Question
3 answers
Dear Colleagues,
Currently, I am making Phospholipase D (PLD) catalyzed transphosphatidylation between Phosphatidylcholine(PC) and Glucosamine in order to get phosphatidyl-glucosamine '( A glycophospholipid). As the phosphatidyl-glucosamine is a new molecule therefore no standard for it, I would like to know how can I quantify the amount of phosphatidyl-glucosamine ?
Your king answers are highly appreciated
Relevant answer
Answer
You could try quantifying the primary amino group using ninhydrin absorbance or o-phthaldialdehyde fluorescence, or other amine detection reagent, with any primary amine-containing compound (e.g. Tris) as a standard.
  • asked a question related to Glucosamine
Question
3 answers
Hi, I have extracted glucosamine from lobsters shell, however my extracts are not soluble in water,Could you please tell me why and give me a suitable way to dissolve them in water?
Thank you
Relevant answer
Answer
Chemically, glucosamine is 2-amino-2-deoxy-alpha-D-glucose. It is a
crystalline colourless and odourless substance with sweet taste. Glucosamine is readily soluble in water with a log P of -2.2 and practically insoluble in ethanol (96%). It has a pKa value of 6.91 at 37°C. When heated above 80oC, it undergoes Maillard reaction and its colour gets faded.
  • asked a question related to Glucosamine
Question
2 answers
Hello,
 I am using EDC and NHS to attach fatty acids (oleic and linoleic acid) onto chitosan (5000 Da) backbone. The final product I am getting is extremely cytotoxic. I am wondering if EDC is the cause of this cytotoxicity, and it is not getting removed properly.
None of the other components I am using for this reaction are toxic to the cells (HEK-293).
 The procedure I am following is a widely used/published protocol. I dissolve chitosan oligosaccharide in distilled water at pH 5. Separately dissolve EDC (1 mol/mol of fatty acid), NHS (1 mol/mol of fatty acid), and fatty acids (0.3 mol/ mol of glucosamine unit of chitosan).
I add the ethanol solution dropwise into the chitosan solution while stirring, and let it stir at room temperature for 12-24 hours. I dialyze this solution against deionized water, repeatedly changing water 2 times a day, upto 4-5 days (protocol says 2 days, I have tried longer). Freeze dry the product, purify with ethanol to remove unreacted fatty acid, freeze dry again.
The final chitosan-fatty acid polymer sample is extremely cytotoxic (5-30% cell viability), while chitosan alone gives me almost 100% cell viability (using MTT assay). I am wondering if it is the EDC/NHS that hasn’t been removed properly by dialysis.
 Can someone suggest something that is wrong with this reaction, with the EDC, or the way I am using the EDC?
Relevant answer
Answer
Thank you Mr. Khan. I took your suggestions into consideration.
However, I figured out that the cytotoxicity was due to residual ethanol and vacuum drying was not sufficient to remove it. I am having a much better outcome with drying in oven at a slightly higher temperature.
  • asked a question related to Glucosamine
Question
1 answer
I'm trying to convert glucosamine into glucose. I've found one paper online explaining how this can be done, but it's over 100 years old, and requires about 10 steps ( CXIX.—The conversion of d -glucosamine into d -glucose ) . Does anyone know if there is a quicker way to do it? The only thing I have to go on right now is that the internal alcohols need to be protected before a deamination agent can be added to the mixture.
Thanks for all the help :)
Relevant answer
Answer
it must be infinitely easier and cheaper to buy glucose.
  • asked a question related to Glucosamine
Question
2 answers
During chitin hydrolysis, non-crystalline glucosamine also obtained whether is it due to caramelisation. If anyone worked on that please give me some references
Relevant answer
Answer
Thank you Rumesh
  • asked a question related to Glucosamine
Question
2 answers
How to generate a SER dipeptide with O-GlcNAc (beta- glucosamine) attachment at side chain with OH group of SER. I have genetared the dipeptide with O-GlcNAc , It shows complete dipeptide when visualized in pymol but not in Chimera (shows GlcNAc is not attached to serine). Thus I can not submit it to charmm-gui for further input file generation. Please Help.
Relevant answer
Answer
Thanks Dileep
Although I have sorted out the problem with CHARMM scripting (using join command to attach OGlcNAc to Ser).
Thanks for your valuable suggestion.
  • asked a question related to Glucosamine
Question
5 answers
Alginate is a water-soluble linear polysaccharide composed of alternating blocks of 1– 4 linked α -L-guluronic and β-D-mannuronic acid residues ,whereas chitosan is a co polymer of D-glucosamine and N-acetyl glucosamine.
They have favourable properties of biocompatibility, biodegradability, pH sensitiveness and muco adhesiveness.
Relevant answer
Answer
OK. What about if we are to encapsulate a mixture of the peptides in form of  hydrolysate which is found to be hydrophyllic upon solubility test.
  • asked a question related to Glucosamine
Question
1 answer
What are the compatible columns of HPLC for analysis of N-acetyl glucosamine?
Relevant answer
Answer
A number of HPLC methods have been reported for the analysis of N-acetylglucosamine. I'll suggest you out some of them and their references in order to find which one will work best for you. Some of these are:
Journal of Food and Drug Analysis 2000. 8(2): 75-83
  • asked a question related to Glucosamine
Question
4 answers
Hello all,
The molecule is a derivative of N-acetylated glucosamine with 3 -OBz groups, the glucosamine is also attached with an aliphatic chain which contains methyl ester.
Please suggest me on substituting the -OBz groups with -OAc groups in one or minimum number of steps. The methyl ester also later to be hydrolyzed.
I tried this in 3 steps, debenzoylation, hydrolysis of ester and acetylation of hydroxyl groups. But the impurity formation is high
Relevant answer
Answer
I think it is a very easy transformation. You can use NaOMe in anhydrous MeOH at ice cold temp to cleave -OBz, neutralizing the basic mixture by HCl or H+ resin, then acetylation. It will not affect the methyl ester, and you can later hydrolyze it by NaOH. 
I have no idea about the impurity, I think it is a very clean reaction. 
  • asked a question related to Glucosamine
Question
3 answers
 I am trying to read about this but I could not get a final answer and clear one 
Relevant answer
Answer
We used lissome to cleave chitosan in a nanoparticulate system in which chitosan contributed to the structure of the nanoparticle. It worked.
  • asked a question related to Glucosamine
Question
1 answer
Leaves dry powder
Relevant answer
Answer
hi 
D-glucosamine-hydrochloride 110nm
i have a research paper on that so u can download this and study 
  • asked a question related to Glucosamine
Question
4 answers
I am trying to develop a HPLC method for
Relevant answer
Answer
If you have polar analytes that not well retained on the C18 or have the same retention time, and you want to separate them well in a single run, I would recommend you to explore the mixed-phase mode or HILIC. The idea is to retain by ion-exchange and elute with pH or salt. Attached file is the example for you to consider.
  • asked a question related to Glucosamine
Question
9 answers
Glucosamine is a chitin-derived mucopolysaccharide and putative disease modifying agent; anti-rheumatic agent.
Relevant answer
Answer
yes, the NSAID effect of Glucosamine sulfate is there, but it is of other class/chondroprotective  when we see it as a NUETRACEUTICAL promotion....But i would like to ask why plain glucosamine sulfate with 1500 mg is not acceptable by the doctors in Grade I osteoarthritis..